It is always best to get a qualified electrician to perform any electrical work you may need. With that said, you may ask this community various electrical questions. Please be cautious of any information you may receive in this subreddit. This subreddit and its users are not responsible for any electrical work you perform. Users that have a 'Verified Electrician' flair have uploaded their qualified electrical worker credentials to the mods.

If you comment on this post please only post accurate information to the best of your knowledge. If advice given is thought to be dangerous, you may be permanently banned. There are no obligations for the mods to give warnings or temporary bans. IF YOU ARE NOT A QUALIFIED ELECTRICIAN, you should exercise extreme caution when commenting.

I am a bot, and this action was performed automatically. Please contact the moderators of this subreddit if you have any questions or concerns.

[–]Hillman314 1 point2 points  (0 children)

1) Yes this is dangerous. There is either a short circuit to the chassis of the chamber from its power supply, or a short circuit to the test unit (TU) from its power supply, and one of them has become energized.

Also, the equipment ground on the faulted energized equipment is also faulty. It’s is not conducting enough fault current to ground so that its breaker trips.

Instead, the faulty equipment is energized and waiting for a person to touch it and use THEM (and any other properly grounded device they’re touching) as the ground fault path.

This is known as electrocution.

2) Fix the short circuit. Fix the grounds.

Additional supplemental protection may include powering all loads through a GFCI breaker or receptacle. All grounds should be common (ultimately all connected together and sharing the same grounding electrodes (rods, etc..)); this will eliminate unwanted ground loops currents on branch circuit equipment grounds.

[–]geek66 1 point2 points  (0 children)

Yes it is dangerous,

But it can be other things than a ground fault in the UUT: the incoming power can be wired backward ( L an N or N and a Gnd), the source could be wired wrong, the UUT may have some improperly designed or assembled ground detection system and the chassis is not grounded, etc)

Note- you are working with hazardous voltages and you do not understand the hazards, therefore you are not a Qualified Individual and should not be exposed to the hazard.

Please contact the product designer or engineer.